Optics - Filter Properties
  • When using laser light sources, filters must have very sharp cutons and cutoffs since there will be many orders of magnitude more scattered laser light than fluorescence


  • Can specify wavelengths that filter must reject to certain tolerance (e.g., reject 488 nm light at 10-6 level: only 0.0001% of incident light at 488 nm gets through)
3
Lecture Goals
  • This lecture is intended to describe the nature and function of optical systems
  • It will describe how filters operate
  • When filters should be used
  • What problems and issues must be taken into consideration


4
Optics - Filter Properties
  • Long pass filters transmit wavelengths above a cut-on wavelength


  • Short pass filters transmit wavelengths below a cut-off wavelength


  • Band pass filters transmit wavelengths in a narrow range around a specified wavelength
    • Band width can be specified


  • Neutral Density filter is a nondiscriminant intensity reducing filter
5
Optics - Filter Properties
  • When a filter is placed at a 45o angle to a light source, light which would have been transmitted by that filter is still transmitted but light that would have been blocked is reflected (at a 90o angle)


  • Used this way, a filter is called a dichroic filter or dichroic mirror
6
Interference and Diffraction: Gratings
  • Diffraction essentially describes a departure from theoretical geometric optics
  • Thus a sharp objet casts an alternating shadow of light and dark “patterns” because of interference
  • Diffraction is the component that limits resolution


7
Interference in Thin Films
  • Small amounts of incident light are reflected at the interface  between two material of different RI
  • Thickness of the material will alter the constructive or destructive interference patterns - increasing or decreasing certain wavelengths
  • Optical filters can thus be created that “interfere” with the normal transmission of light
8
Optical filters
  • Interference filters: Dichroic, Dielectric, reflective filters…….reflect the unwanted wavelengths
  • Absorptive filters: Colour glass filters…..absorb the unwanted wavelengths
9
Interference filters
  • They are composed of transparent glass or quartz substrate on which multiple thin layers of dielectric material, sometimes separated by spacer layers .


  • Permit great selectivity.
10
 
11
Standard Long Pass Filters
12
Dichroics
  • They used to direct light in different spectral region to different detectors.
  • They are interference filters , long pass or short pass.
  • "dichroic" Di- is Greek for two, and -chroic is Greek for color - from Greek dikhroos, bicolored
13
Optical Filters
14
Dichroic Filters
15
Construction of Filters
16
Transmission determination
  • Constructive and destructive interference occurs between reflections from various layers
  • Transmission determined by :
    • thickness of the dielectric layers
    • number of these layers
    • angle of incidence light on the filters
17
Absorptive filters
  • Such as coloured glass filters which absorb unwanted light.
  • Consist of dye molecules uniformly suspended in glass or plastic.
  • Remove much more of the unwanted light than do the interference filters
  • Will often fluoresce (not good!)
18
Filters transmission
  • Bandpass filters: characterized by there
  • T max and (the Full Width at Half Maximum) FWHM


  • Notch filters are band pass filters in the upside down position


  • Long pass and Short pass filters: characterized by their T max and cuton, cutoff wavelength.
19
 
20
Interference filters advantages
  • They can be used as reflectors in two and three color analysis.
  • They usually do not themselves produce fluorescence.
  • They are available in short pass versions.
  • They are excellent as primary barrier filters.
21
Interference filters: disadvantages
  • Lower blocking properties
  • Reduced passing properties
  • Their reflecting and passing properties are not absolute, this should be considered while dealing with multiple wavelengths
22
Absorbance filters: advantages
  • They are inexpensive.
  • They have very good blocking properties.
  • They have very good transmission properties.
23
Absorbance filters: disadvantages
  • They can only pass long wavelengths
  •      ( hence, can only block short   wavelength)
  •  Since they are made of solution of dye and glass, they can themselves produce fluorescence.
24
Neutral density filters (N.D)
  • Attenuation of the light without discrimination of the wavelength.
  • N.D filters could be reflective or absorptive type.
  • They are partially silvered mirrors.
25
Beam splitters
  • Absorptive N.D filters can not be used here; simply because of the heat, they will melt.
  • Common cover slips can be used as beamsplitters if small portion of the light is wanted, up to 5%
26
Measuring Filter Properties
  • Filters must be measured at the angle they are going to be used
  • filters placed at 90o have different properties when they are placed at 45o
27
Short pass and long pass filters
28
Optical filter evaluation
29
Optical filter evaluation
30
Light loss in dichroics
  • Reducing reliance on the in line arrangement PMTs
  • Placing a second fluorescence collection lens at 180o from the first one (this is more difficult in most instruments)
31
Light loss by optics
  • The thicker the glass the less light transmitted.
  • Problems with glass - UV light will not pass
  • In UV light system use minimum optics.


32
Light loss by optics
  • Glass can absorb UV light and can fluoresce when illuminated at that wavelength.
  • For excitation > 450nm, you can use glass filters, < 450nm use quartz or silica filters.
  • Plastic optical filters are unsatisfactory


33
Optical filters evaluation
  • Use a population of appropriately stained particles and identify which filters give the maximum signal.
  • Spectrofluorometer amd spectrophotometers can be used as tools for assessment of optical filters.
34
Issue to Note
  • Problems with filters are more likely due to using the wrong filters
  • Filters degrade overtime, so they have to be changed eventually
  • Buy high quality filters, not cheap ones
35
Hints on filters
  • To obtain acceptable blocking of the light outside the pass band, most interference filters incorporate some absorptive elements as well as dielectric layers
36
More hints...
  • You have to be careful while using short pass filters, specially with short wavelength, because of the transmission ability of these filters for long wavelengths (they behave like notch filters). If you have long red/near IR signals they will pass
37
In general
  • Use the least number of filters necessary to reduce signal loss
  • Absorption result in conversion of light into  heat. Thus, laser beams hitting colour glass filters may destroy these filters .
  • Filters have a finite lifetime.
38
Practical consideration
  • In measuring weak fluorescence, we usually use field stop and interference filters behind the field stop to remove the stray light.
  • The shiny part (mirror side) of the filter should face the light source (collection lens)
39
Fiber optics & optical waveguides
  • Fiber optics and other optical waveguides operate by total internal reflection
  • problems with stray light, low NA of fibers, thus low sensitivity, light collection difficult
40
Optics for forward scatter
41
Lecture Summary
  • At the conclusion of this lecture the student should understand:
  • Field stops and obscuration bars are necessary in systems where air or round capillaries are used
  • Appropriate optical filters must be placed in combinations
  • Filters degrade over time and should be checked
  • The least number of filters should be used in a system
  • Forward angle scatter is frequently collected using a diode detector