BMS 631 - LECTURE 5
Flow Cytometry: Theory

J.Paul Robinson
Professor of Immunopharmacology & Biomedical Engineering
Purdue University
2
Illumination Sources
  • Lamps
      • Xenon-Mercury
      • Mercury

  • Lasers
      • Argon Ion (Ar)
      • Krypton (Kr)
      • Helium Neon (He-Ne)
      • Helium Cadmium (He-Cd)
      • YAG
3
Optics - Light Sources
 Epilumination in Flow Cytometers
  • Arc-lamps
    • provide mixture of wavelengths that must be filtered to select desired wavelengths
    • provide milliwatts of light
    • inexpensive, air-cooled units
    • provide incoherent light
4
Mercury Arc Lamps
5
Arc Lamp Excitation Spectra
6
Optics - Optical Channels
  • An optical channel is a path that light can follow from the illuminated volume to a detector
  • Optical elements provide separation of channels and wavelength selection
7
Spot Illumination - Lasers
  • Advantages are that the pathway is easier to define (you know where the light is going !!)
  • It is usually monochromatic light so excitation filters are not needed
  • Brighter source of light than arc lamps (higher radiance)
  • Spot size (d) can be calculated by formula
    • d=1.27(lF/D)   where D is the beam diameter in mm and F is the focal distance from the lens
  • For a 125 mm focal length spherical lens at 515 nm is 55 um and 61 um at 458 nm
8
Lasers
  • Coherent Enterprise laser - UV-visible
  • Air cooled laser (Argon)
9
Laser Power & Noise
Light Amplification by Stimulated Emission of Radiation
  • Laser light is coherent and monochromatic (same frequency and wavelength)
  • this means the emitted radiation is in phase with and propagating in the same direction as the stimulating radiation
  • ION lasers  use electromagnetic energy to produce and confine the ionized gas plasma which serves as the lasing medium.
  • Lasers can be continuous wave (CW) or pulsed (where flashlamps provide the pulse)
  • Laser efficiency is variable - argon ion lasers are about 0.01% efficient (1 W needs 10KW power)
10
Lasers
11
Argon & Krypton Lasers
12
Dye Lasers
  • Dye lasers use a source laser known as the pump laser to excite another laser known as the dye laser.
  • The dye laser consists of a flowing dye which exhibits desirable properties such as excitation and emission.
  • The lasing medium is a fluorescent dye (e.g. Rhodamine 6G) which is dissolved in an organic solvent such as ethanol or ethylene glycol
  • The laser can be tuned, usually by a rotatable filter or prism
  • The dye must be circulated and cooled to prevent it being bleached or over-heated
13
Helium-Neon Lasers
  • He-Ne - low power, no cooling needed
  • Cheap, mostly emit red light at 633 nm
  • Generally 0.1 W to 50 mW power
  • Lines available   include green (543nm) and red 633 nm,  594nm or 611 nm.


14
Helium-Cadmium Lasers
  • He-Cd laser
  • 5-200mW power usually at 325 nm (UV) or 441 nm (blue)
  • Wall power, air cooled
  • Uses cadium vapor as the lasing medium
  • Major problem is noise (plasma noise between 300-400 kHz)
  • RMS noise mostly about 1.5%
  • Good for ratio measurements (pH or calcium) because power fluctuations don’t matter here – these lasers do have power fluctuation problems eventually.



15
Diode Lasers
  • Small, efficient, cheap
  • Only red wavelengths available at reasonable prices (blue works, but still problems)
  • Mostly made of Gallium aluminum arsenide (GaAlAs)
  • Emission ratio is varied by changing the ration of gallium to aluminum in the semiconductor
  • Main use is CD players (now 2 in every household!! One in the stereo and one in the computer! And maybe one in the laser printer!)
  • Biggest problem is not power - but lack of fluorescent probes to be excited at 650-900 nm
  • Problem is poor beam profiles for diode lasers
  • Noise levels are generally 0.05% or less  compared to 1% for air cooled argon and .02% with water cooled argon lasers




16
Solid State Lasers
  • Neodynymium-YAG (Yttrium aluminum garnet) lasers
  • Lasing medium is a solid rod of crystalline material pumped by a flashlamp or a diode laser
  • 100s mWs at 1064 nm
  • can be doubled or tripled to produce 532 nm  or 355 nm
  • Noisy - and still reasonably expensive (particularly the double and tripled versions)


17
Lasers Hazards
  • Laser light is very dangerous and should be treated as a significant hazard
  • Water cooled lasers have additional hazards in that they require high current and voltage in addition to the water hazard
  • Dye lasers use dyes that can be potentially carcinogenic
18
Summary so far….
  • Arc lamps are useful for flow cytometry because of low cost and wide spectral characteristics
  • Arc lamps require more complex optical trains
  • Lasers provide light at high radiance
  • Lasers are essentially monochromatic, coherent
  • Lasers represent a significant hazard


19
Goals of Light Collection
  • Maximum signal, minimum noise
  • Maximum area of collection
  • Inexpensive system if possible
  • Easy alignment
  • Reduced heat generation
  • Reduced power requirement


20
Optical Collection systems
21
Objectives
  • 1.3 NA objective
22
Field stops & obscuration bars
  • Obscuration bar is placed along the path of the illuminating beam
  • It blocks the direct light but allows the fluorescence signal (which is going in all directions)
  • In a capillary or cuvet system, a field stop which is placed in the image plane achieves the same result
23
Optical translators
24
The point of a good optical system is to obtain a good Signal Vs Noise
  • Good optical filters
  • Remove as much excitation signal as possible
  • Collect as much fluorescence as possible (use concave spherical mirrors)
25
Spectral Selection
(Next lecture)
  • Monochromators Vs Filters
  • Filters are reasonably inexpensive
26
Lecture Summary
  • After completing this lecture you should understand:
  • Excitation light sources and their properties
  • Each light source has unique utility
  • Optical components together with light source creates an optical system
  • The general nature of optical systems in typical cytometers